We have a cell temperature exceeded alarm. Does this mean the cell heater needs to be replaced?

Gary Brewer

ABB's answer

No necessarily.  If the ambient temperature exceeds 100F then it can also cause this alarm.  Check the ambient temperature of the analyzer and if it is below 100F then the cell heater is probably bad.